A measure is not detected
I was going to control Pulse Generators (WF1945B) of NF company,
It was not detected in MAX.
The interface is GPIB.
Is it necessary to do what it is to let you detect it?
An answer thanking you in advance.
Message 1 of 13
(4,075 Views)
check to make sure that GPIB is enabled on the instrument. Check that the address is unique and not the same as the controller. Check that you can detect another instrument.
Message 2 of 13
(4,074 Views)
As for the upper sentence, what is the address of the controller?
GPIB is recognizable more.
However, an error is given at the time of scan of the measure.
Therefore the control of the pulse generator is not possible.
Message 3 of 13
(4,070 Views)
The controller is usually 0 and you can see that in MAX.
if you are getting an error, provide the code. We are not mind readers.
Message 4 of 13
(4,058 Views)
I'm sorry.
There seemed to be little information.
Besides , if there is information necessary , I ask .
I write down the error code below.
-1073807346
Message 5 of 13
(4,054 Views)
I would expect that VISA error if MAX fails to find an instrument. MAX does the scan and finds nothing and does not report an error?
Message 6 of 13
(4,021 Views)
I am sorry that a reply is late.
It is the street.
The message of the error is displayed when I scan a measure in MAX without finding anything.
Message 7 of 13
(3,999 Views)
Sorry, I don't understand what you mean.
Message 8 of 13
(3,988 Views)
I am sorry in poor English.
The message of the error is displayed when I let you scan a measure in MAX.
Message 9 of 13
(3,977 Views)
That error code does NOT come from doing a scan for instruments in MAX. It comes from a VISA function and VISA is not used in the scan for instruments. If the scan for instruments does not return anything, it is a waste of time to try any VISA function. The scan for instruments will first have to run and find something.
Message 10 of 13
(3,957 Views)
