Using LabView version 7.1, VISA 4.4.1, and communicating using RS-232 to an Agilent 34970a with a 34903 20-channel actuator card.

 

On using the example VIs and drivers downloaded from the Agilent website, I am able to send commands to open and close switches on the 34903 card without issue most of the time.  However, I have discovered that, whether using the 'open' or 'close' command, that the length of the command string matters.

 

If the command string, which will look something like: ROUT:OPEN (@101,102,103), has a total number of characters of 80 or less, then there's no problem. If the total number of characters is greater than 80 (like if I ask it to close or open all 20 channels on the card), then I get Agilent erros 521 (Input buffer overflow) and -103 (Invalid separator).  

 

It seems as though somewhere in the path of generating the byte string to send to the instrument that something like a CRLF character is being inserted after the old standard 80 bytes.  But a probe on the string which is being sent to Visa Write doesn't reveal any such thing.  Any thoughts on where this might be happening (within VISA?)  and what I might do to correct it?

 

 

This is odd because I'm able to write a much longer configuration command string (many hundreds of characters) when configuring an 34901 card for temperature measurement.

 

Your suggestions are greatly apprecaited,

Troy.

0 Kudos
Message 1 of 12
(8,500 Views)

The termination character is required by the instrument to process the command and you should not get rid of it. It is being sent after every command.

 

You said you got the driver from Agilent. Have you tried the driver from NI?

0 Kudos
Message 2 of 12
(8,497 Views)
I'm afraid I mis-wrote.  The drivers I'm using are, in fact, from NI.  They're part of the standard bundle in the hp34970a.zip file.
0 Kudos
Message 3 of 12
(8,489 Views)
I'm not sure why you would be getting an error. I never used the instrument with a serial connection and I don't know if there is a serial buffer that can be configured. I also no longer have any of the instruments to experiment with. What you could try is sending the command in MAX to see if there is a problem with the way the driver is creating the string. If that also generates an error, try configuring the instrument with the Agilent utility program that comes with the instrument. If that also generates an error, you should be contacting Agilent.
0 Kudos
Message 4 of 12
(8,485 Views)

Dennis, Thanks for your attempts to help

It turns out that I was making an operator error (of the somewhat embarassing kind) and that now the issue has been solved.  😉

 

Thanks again,

Troy

0 Kudos
Message 5 of 12
(8,470 Views)

I'm experiencing the same issue, I've only noticed this problem after switching over to a serial server instead of a direct connection.  Could you give me information on the "operator error" that you determined so that I can check if I'm doing the same?

 

Thanks,

Randy

0 Kudos
Message 6 of 12
(8,297 Views)

Hi Redrabbit,

What do you mean by a serial server? You could take a look at a spy capture to try and see what is different between the two cases. Let me know if you’re still having problems.

Matt
Applications Engineer
National Instruments
0 Kudos
Message 7 of 12
(8,284 Views)

By serial server I mean a device which connects to the LAN and provides access to serial devices via virtual com ports (mapped over the network).  I actually figured out what I was doing differently vs. the directly connected instance of this application, I had set everything up with no flow control whereas in the direct connect it was setup as XON/XOFF.  Not sure why I can't make "no flow" work (all devices are set appropriately) but I guess I'll shrug and move on.

 

Thanks,

Randy 

0 Kudos
Message 8 of 12
(8,281 Views)

Hello,I recently started to learn LabVIEW,and I've met with the problem you used to met.Now I am eager to know what kind of "operator error" you meaned about .Could you tell me about it?

 

thanks 

jiacheng

帖子被emptytiger在 04-09-2010 10:36 PM
时编辑过了
0 Kudos
Message 9 of 12
(7,566 Views)

Hello,I've read the posts you wrote about seven months ago.I now met the same error .Could you tell me how did you eliminate the "error 521:Input buffer Overflow".

thanks

jiacheng

0 Kudos
Message 10 of 12