Error -1073807202 Missing Code Library in VISA Open
I'am trying to communicate with ANRITSU MS9710B Optical Spectrum Analyzer via GPIB. Anritsu provides VISA-based VIs for all functions. When trying to initialize the instrument (by use of VISA Open), I'm getting the following error: -1073807202, that is "a code library required by VISA could not be located or loaded".
I can communicate with the instrument using the Measurement and Automation Explorer, and have recently installed the newest version of NI-488 driver (Version 2.2 for GPIB-USB-B).
I'm using Labview 8.0.
Where can I find the missing code library?
Thanks in advance!!!
Freddy
(5,568 Views)
You need NI-VISA which is not the same thing as NI-488.2. You should have it on the LabVIEW driver CD, the CD that came with the GPIB card, or here.
(5,566 Views)
Accepted by topic author freddybee
hm...yes indeed! I had actually installed NI-VISA, but MAX indicates that it was not properly done. I'll get rid of  it and re-install the VISA drivers,...that might just be it! I'll let you know....in the meantime:
Thanks for your quick answer!
(5,560 Views)
...cool, it works... THANKS
(5,555 Views)
The solution is not your thank you. You can change the marked solution by going to 'Options' right above the message body.
(5,552 Views)
