When I do connect everything and enter in the address of the TX PS2520G power supply it gives me an error in the OLD Visa Open,
Error -1073807195 occurred at Old VISA Open in TKPS252XG Initialize.vi
Possible reason(s):
VISA: (Hex 0xBFFF00A5) The interface type is valid, but the specified interface number is not configured.
07-10-2006 07:32 PM
07-11-2006 04:18 PM
07-11-2006 04:39 PM
Yes.
I'm curious though about the driver that you downloaded. The 8.0 version of the driver does not have a 'Old VISA Open'. The 7.0 and 6.0 versions do and if you want to use the driver, the 'Old VISA Open' will have to be replaced. If you just download the 8.0 driver, your life will be a bit easier. The replacement is not difficult but it's a step that you can easily avoid.
With an NI GPIB-PCI card setup and a successful scan for instruments, when you open the driver (8.0 or modified one), you can click on the right side of the VISA Resource Name control and get a list of all VISA resources on your pc. VISA resources will include com and parallel ports so you just want to select the GPIB resource that corresponds to the instrument you have connected.
07-12-2006 10:18 AM
07-12-2006 04:44 PM
07-12-2006 04:53 PM
07-12-2006 05:55 PM
07-12-2006 05:59 PM
GPIB is a shared bus that you can theoretically connect up to 32 devices simultaneously. I say theoretically because there are physical limitations on both bus loading and cable length before things go a bit south.
I typically have anywhere from 3 to 9 instruments connected to one bus with no problems.