Thank you for posting to the NI Discussion Forums.
The error that you are receiving indicates that the software circular buffer of the counter task is full and therefore previous samples were overwritten causing a loss of data. The software buffer size is directly dependant on the mode of operation as well as the acquisition rate. The actual buffer size can be found on page 2-8 of the NI 660x User Manual, which can be found here:
https://www.ni.com/docs/en-US/bundle/ni-660x-feature/resource/372119c.pdf
So, part of the troubleshooting process is to find out what rate you are sampling at and if you are doing finite or continuous sampling on the counter task. This will help us determine what size buffer you currently have and thus if it is likely that we are overflowing this buffer.
To prevent the overflow from happening so soon, we can increase the size of the buffer. This can be done using the DAQmx Buffer Property node which is available on the DAQmx functions palette>>DAQmx Advanced Task Options. See if increasing this value improves the performance of your application.
Let us know your findings and we will be happy to help you further.
Have a great day!
Brian F
Applications Engineer
National Instruments
05-09-2007 02:33 PM
Thanks for the help Brian,
We are using Continuous Sampling, and I've been controlling the Buffer Size using the Samples per Channel input on the DAQmx Timing (Sample Clock).vi . Will this do the same thing as altering the buffer property node?
Because the Photons arrive randomly, it is difficult to state an actual rate at which they are arriving. We are seeing about 5k-40k counts per second, but they are randomly spaced. Could a possible cause of this error just be that two photons happen to arrive too closely?
I noticed on the link you referred that finite operation seems to allow for much higher sampling rates. We have been gating our data aquisition by triggering initiation of collection on the rising edge of a pulse and then running in continuous sampling for a little longer than our gate period, then truncating any samples that are timestamped as having arrived outside of our gate period. Would it possibly eliminate our error to find a similar way of accomplishing this with finite operation, even though we don't know exactly how many samples will arrive in our gate window?
Attatched is a copy of our code
Thanks again for the help
LJ
05-09-2007 03:39 PM
One other thing with regards to our signal that I failed to mention first. The TTL pulses that the photon detectors produce when a photon arrives are 30 ns long (2.5-5V), but there is 50 ns of dead time between the pulses. This means that the detectors can't send pulses spaced less than about 80 ns apart. So the quickest signal that the photon detectors could possibly generate, assuming photons were arriving at that rate, would be about 12.5 MHz. Once again, the photon arrivals are random, but it is possible that we could get a pair once in a while that are this close together, even though our count rate is MUCH less than 12.5 MS/sec
LJ
05-10-2007 05:39 PM
Hello.
This error results from the fact that the counter FIFO size is only one sample, so that error -200141 is thrown if two consecutive samples infringe the specs. For example, a 10,000 sample buffer should handle 1.6 MS/s according to spec, which means that two consecutive signal edges must have a time interval of at least 625 ns. So, if you receive two photons within 625 ns, the 1 sample counter FIFO will be overflowed and data will be lost. The best solution for this application is to buy a card with a larger counter FIFO. The 6210 is probably the best card for this as it features a 1,023 sample FIFO instead of a 1 sample FIFO.
Let me know if I am explaining this clearly. Have a great day!
Brian F
Applications Engineer
National Instruments
05-11-2007 07:10 AM
Isn't the 1023-sample FIFO for analog tasks? I thought that all the M-series counter boards were limited to a 2-sample FIFO, just like the 6602...
Previous threads on similar apps seem to confirm that finite acquisition can be a better choice due to less DMA-related overhead. I can't seem to find the exact thread right now, but search on terms like counter, finite, fifo, dma, 200141, etc. One of the guys from NI had a real good explanation about the nature of why finite acq can sustain higher sampling rates than continuous acq. I don't remember who it was just now, but two likely candidates that come to mind are reddog and gus. I also think that the thread was sometime in the last several months.
- Kevin P.
05-14-2007
10:01 AM
- last edited on
02-12-2025
11:40 AM
by
Content Cleaner
Hello.
You are correct in asserting that normal M series boards have a fairly small FIFO. However, the 6210 is a member of our Bus Powered M Series devices, which have a much larger counter FIFO.
Brian F
Applications Engineer
National Instruments
05-14-2007 10:39 AM
Thanks guys,
The information you've given has been very helpful. Is there a way to tell the 6602 board to disregard the error message I can still live if I lose one of two closely spaced signals, but the big problem has also been that whenever the error gets tripped, the array that I'm building (basically the code above just reads the buffered data in a while loop and appends it onto a growing array with each iteration) has a whole bunch of zeros appended onto it at the end. For example, the counter might be running just fine and at about, say, 4000 counts the error gets tripped. Suddenly my array is no longer 4000 elements but 16000 elements, and the last 12000 are all zeros. Is this just a problem with my code, or is this some problem that's inherent in the error itself? If I can tell the counter to just keep running like normal in this operation and disregard one of the signals, it would solve my problem. At least it would be a preferable solution to buying something to filter out closely spaced signals.
If not, would the M-series 6210 work well with this application? Basically I chose the 6602 because it has 8 counters and an 80 MHz internal time bas with which to timestamp. If the 6210 only has 2 counters, I suppose it would still work, but would I still be able to timestamp in the same manner?
Thanks again for the help, everyone
LJ