Sean's suggestion about buffered edge counting should work well and will be particularly handy if you want to know "at a glance" the cumulative total # of TTL pulses that have come in. You can find the pulses per 1 msec bin by doing a simple finite difference. Everything will be hw-timed, and you won't miss any pulses between bins.
Another possibility is to do buffered period measurement with the DAQmx units set to "Ticks." This would directly store the # of pulses during each 1 msec bin time.
Either way, you'll probably need to set the "Duplicate Count Prevention" property (buried down deep in the DAQmx Channel property node). I
think the right setting is "True", but you can always try both True and False to see which works properly. One setting will actually store 0 values during 1 msec bins with no incoming TTL pulses, the other setting will simply ignore such intervals and only store a value during the intervals where there are 1+ pulses.
Finally a comment about your mention of "Nyquist criteria" in your original post. For most counter tasks, you wouldn't generally consider Nyquist criteria the way you would in an analog task. Consider your app: your 1 msec bins are the same as sampling at 1 kHz. However, it's apparent that you expect several TTL pulses to occur within some of those 1 msec bin times. So your signal can come in as fast as (perhaps) 5 kHz or more. If you automatically think of Nyquist, you might too quickly decide to sample at >10 kHz, which is both unnecessary and actually unhelpful for your app.
-Kevin P.