Hi Uven,
Thanks for contacting National Instruments. I understand why you might want to use only one channel for this acquisition; however I believe that the result you are looking for will require two input channels. This is because each channel on the TC-120 is a differential input. The temperature that is received on each channel is a measurement of the difference in potential of the two metals of the thermocouple. Therefore, both terminals of the input channel are necessary to obtain the temperature from each side of your system.
Your current configuration seems to be the best way to obtain the overall temperature difference. I hope this helps. Let me know if there is anything else I can help clarify. Have a great day!
Jason W.
01-01-2008 02:00 PM
01-02-2008
01:49 PM
- last edited on
03-13-2025
03:33 PM
by
Content Cleaner
Hi Uven,
The cFP-TC-120 is a specialized module for measurements using standard J, K, T, N, R, S, E, and B thermocouples. With all of the built in functionality, the values that are obtained from the cFP-TC-120 are reliable, accurate temperature or millivolt measurements. The values are thus bounded and scaled based upon the 'Range' selection in the 'Channel Configuration' page. The differential voltage measurements that I believe you are looking to obtain can be achieved with a different type of module, such as the cFP-AI-118. FieldPoint modules are typically designed for more specific purposes, powered by our FieldPoint driver, as opposed to our PCI/PXI DAQ devices that are controlled by DAQmx.