Switch Hardware and Software
How to configure hardware for 120 different analog input signals?
For one of the application we need to choose a hardware(switch matrix) so that we can switch over 120 different inputs.
We are using Digital Multimeter for the measurements of these inputs.
But i stucked on how to select good combination of switch matrix.
Please give me advice.
(7,340 Views)
Hi adat,
I advise establishing a set of switching requirements for your application.  I've provided a list of specifications and features that you can use in determining the requirements for your application:
Bus Type (PXI or SCXI)?
OS Support (Windows XP, Vista, Linux, Real-Time, etc.)?
Switch Type (External Relay Driver, General Purpose, Matrix, Multiplexer, RF, etc.)? - Matrix or Multiplexer
Relay Type (Armature, Reed, FET, Solid State, etc.)?
Connection Speed (Scan Rate, Actuation Time)?
Measurements to Perform (Type, Accuracy, Resolution, etc.)?
Maximum/Minimum Voltage?
Maximum/Minimum Current?
Maximum Power?
Maximum Bandwidth?
Crosspoint/Channel Density? - 120 Channels
Maximum Number of Chassis Slots?
Module Mating Connector Type?
Relay Lifetime?
Answer as  many of these questions as you can, and we'll be able to assist you further.
Thanks and best regards!
Chad Erickson
Switch Product Support Engineer
NI - USA
(7,330 Views)
Hello Chad,
Thank you very much for your valuable reply.
Please find remarks in Blue for more info.
PXI.
Windows XP.
Multiplexer.
Electromechanical relays,latches.
100 cycles/s & 10ms.
Need to measure voltage drop across resistive loads by using DMM.
Maximum: 0-100V, Minimum: 0-12V.
Maximum: 0-100mA Minimum:0-1.2mA.
Maximum Power?
10W Maximum.
Maximum Bandwidth?
Not sure.
10 maximum slots.
Please advice for mating connector. we already chosed NI TB-2636 terminal block.
Relay Lifetime?
1000000 cycles.
Waiting for you reply friend!
Thanks & Regards,
Adat.
(7,311 Views)
Hi Adat,
I recommend the NI PXI-2575.  If for some reason this module is unusable, the NI PXI-2527, PXI-2529, and PXI-2576 could also be suitable for this application.  Here are the specifications for the NI PXI-2575:
Added remarks in Red.
PXI.
PXI.
Windows XP.
Windows XP.
Multiplexer.
Topology can be configured by software to be a 1-wire 196 × 1, 2-wire 95 × 1, or 2-wire 98 × 1 Multiplexer.
Electromechanical, latching relays.
Scan Rate Benchmark w/ NI PXI-4071 DMM:
105 cycles/s (4.5 digits of resolution, 10 V range), 100 cycles/s (5.5 digits of resolution, 10 V range)
DevZone Document with benchmarks attached:
NI PXI-4071 - DC Voltage - DMM/Switch Handshaking Scan Rates
http://zone.ni.com/devzone/cda/tut/p/id/8137
Actuation Time: 1 ms (typical), 3.4 ms (max)
Recommended for switching voltage.
Maximum:100 V CAT I, Minimum: 10 μA @ 10 mV.
Maximum: 1 A, Minimum: 10 μA @ 10 mV.
Maximum Power?
10W Maximum.
30 W, 37.5 VA.
Maximum Bandwidth?
Not sure.
20 MHz.
196 (1-wire) or 98 (2-wire).
10 maximum slots.
Each module fits in 1 PXI slot.  If a 1-wire configuration is used, only 1 slot is required.
Advised connectivity is either:
a) the LFH200 to Bare Wire Cable Accessory, or
b) the LFH200 to 4 50-Pin DSUB -S (CH-CH Twisted) Cable Accessory and 4 x TBX-50 Connector Blocks.
Relay Lifetime?
1000000 cycles.
2500000 cycles (10 VDC, 100 mADC resistive).
Hope this helps!
Chad Erickson
NI - USA
(7,305 Views)