PXI

cancel
Showing results for 
Search instead for 
Did you mean: 

Is it possible to record data directly from PXI-5112 scope card through PXI bus to SCSI RAID array (connected to PXI-8210)?

Colleagues,

Is it possible to record data directly from PXI-5112 scope card through PXI bus to SCSI RAID array (connected to PXI-8210)?
Which will be the maximum transfer rate for continuous data recording?

Thank you,
Sergey
Sergey
0 Kudos
Message 1 of 2
(5,117 Views)
Sergey,
The PXI-8210 can connect to any SCSI 2 compliant device. If the RAID controller is SCSI 2 and appears just like a hard drive in the operating system, then you can send data directly to the RAID array. The problem is that the driver for the PXI-5112 does not yet support continuous acquisition. The on-board 16MB or 32MB buffer stores the data until the entire acquisition is completed. Once the acquisition complete, all of the data is transferred from the on-board buffer to the hard drive. After that happens, the NI 5112 is ready for another acquisition.

Best Regards,
Jace Curtis
NI Applications Engineering
0 Kudos
Message 2 of 2
(5,117 Views)