Multifunction DAQ
usb-6008 to measure/display force
Hello all,
         I am trying to perform some force measurements for a personal project. I have limited labview experience. I have an omega lcl-040 load cell and am acquiring data using the NI usb-6008. What I would like to do is measure the voltage coming from the load cell and then display the voltage signal, resulting force (based on the force-voltage relationship), and the mass in labview. I have run some tests using some pre-loaded VIs and VIs I found online, but want to create my own VI. What I would like to do is be able to set a load on the sensor, read the signal for about 10 seconds or so, then have labview take the average voltage from the readings taken in that time. One issue I am having is that the load cell puts out approximately 2mV when unloaded and I was wondering if there was a way to offset this so that the readings will start at zero when unloaded. If anyone has any suggestions on where I should start it would be greatly appreciated.
Aaron
(3,627 Views)
Unfortunately load cells are not complete precision instruments. They drift. The have offsets. They may have hysteresis, although it is low in good load cells. And the have a small output signal riding on a large common mode signal.  Other than that, they are wonderful.
From the Omega specification the zero balance is 0.3 mV/V. So a 2 mV offset is near the maximum with 5 V excitation and well below maximum for 12 V excitation.
To take advantage of the limited capabilites of the USB-6008 you should add a signal conditioning circuit between the load cell and the USB-6008. It should include a differential amplifier with >60 dB CMRR, gain of 100-500, and an offset adjustment. Depending on your application some filtering may be appropriate.
Lynn
(3,622 Views)
What is the CMRR of the analogue inputs in the usb-6008 board?
(835 Views)
