01-20-2020 08:06 AM
I have a PCI-6225 connected to 2 SCB-68 with 9 pressure transducers wired into "Dev 1". I run this with a Labview 2014 program to collect the pressure transducer data and put it in a file on a computer with Windows 7. I've been running this for about 5 years. I recently moved the PCI-6225 board to another computer with Windows 10 and Labview 2019. I copied over my vi files but when I ran the program, I didn't get the correct readings from the pressure transducers. I moved the board back to the original computer but I now have the same problem on that computer as well. When I look at the board with NI MAX, it seems like it's working fine and the self-test checks out. When I run a test panel for one of the transducer channels, the reading just climbs until it reads 10 V. Any suggestions on how I can fix this?
01-20-2020 08:25 AM
Sounds like a loose/disconnected wire. Double check your wiring in the connector blocks.