my application is to acquire signals from 2 microphone sensors that are connected to two different 6009 daqs.i want to synchronise them in such a way that i can perform further signal processing.
03-07-2015 09:26 AM
Somewhat. The analog inputs of the USB-6009 can use an external digital trigger connected to PFI0. However, the sampling clocks will still be independent. The timing accuracy is specified at 100 ppm.
In the worst case where one device was 100 ppm above nominal and the other 100 ppm below and you sampled at 48 kS/s for one second, the last samples on each device would be 0.2 ms apart. In that time sound in air will travel about 70 mm.
One way to at least partially compensate for this is to record a known identical signal on both devices and see how the data sets differ.
Lynn
03-08-2015 01:36 AM
i am new to labview,can you please elaborate what exactly i shud do to get them in sync