cDAQ-9178 with NI 9221 connected, signal issues

Solved!
Go to solution

I have a cDAQ-9178 with NI 9221 connected taking 3 pressure readings.

 

If a known pressure is applied to any channel, I get 5V 

If the same pressure is applied to 2 channels, I get 2.5V

 

Everything is wired correctly.

 

Could someone point me in the right direction for next steps troubleshooting please? 

0 Kudos
Message 1 of 6
(731 Views)

It sounds like a ghosting issue. Try How Do I Eliminate Ghosting from My Measurements?

0 Kudos
Message 2 of 6
(641 Views)

@ZYOng wrote:

It sounds like a ghosting issue. Try How Do I Eliminate Ghosting from My Measurements?


They're not adjacent, the channels are currently separated (using 0, 2 and 4) with 1 and 3 being left blank. It was previously connected to 0,1,2, but I spaced them to try and prevent any ghosting. 

Setup is 3 x 1Bar pressure transducers on a 10V supply, max 5V output. 

Sample rate is 1Hz Sample size 1

0 Kudos
Message 3 of 6
(633 Views)

Could you please explain the connections and how you apply the signal?

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 4 of 6
(631 Views)

@santo_13 wrote:

Could you please explain the connections and how you apply the signal?


Each pressure transducer has a 4 pin connector. 

1&2 are 9-32Vdc supply voltage pins.

3 is a +ve signal output

4 in earth. 

 

1+2 connected to uninterruptible power supply currently set to 10V

3 is connected to ai0 (and ai2 & ai4 respectively) 

4 on all three are commoned into the ai9 NI9221 ground connector

0 Kudos
Message 5 of 6
(611 Views)
Solution
Accepted by topic author LVIEWPQ

So it would appear that it's a fault with the the grounding on the NI 9221.

When connected to the supply ground, instead of the NI 9221, the ghosting disappears. 

Message 6 of 6
(574 Views)