Multifunction DAQ
RS-232 for reading analog voltage
I am trying to read an analog input of 0-5V from my flow meter, Omega FMA-1605A. I connected the communication wires to the RS-232 port on my USB to RS-232 adapter that is connected to the computerw. But when I try to set up a DAQ read task, it says "No supported devices found."
What can I do to fix this? Should I try using the PXI-6239 card I am using to read analog current signals to read the analog voltage?
Thank you!
Amanda
(4,274 Views)
DAQ has absolutely nothing to do with serial communication. You should be using the VISA functions. There are some shipping examples that you can use for a starting point. I would recommend you first use a terminal emulation program such as hyperterminal, putty, etc. to debug the communication and verify com port settings, commands, and cable type.
p.s. In the future, use the Instrument Control board.
(4,271 Views)
If you want to use a DAQmx task, then you have to use an NI DAQ device.  A simple one like this might work for you: USB-6000
GCentral
(4,236 Views)
You can use the NI-VISA driver to communicate with the Omega device via RS232. Serial communication works by sending device-specific commands. The commands are likely listed in the Omega device's manual. The PXI-6239 is an analog input module, so you can wire signals into the channels. I'm not familiar with the Omega device, so I don't know is it has analog output channels to input to the analog input channels on the PXI board. Most likely you will be using NI-VISA to send and receive messages with the Omega device.
Taylor B.
(4,224 Views)
