07-23-2014 10:36 AM
Hi!
I am using a PCI 6040 E connected to a BNC 2010 to measure 3 different voltages to measure the output voltages of a quadrant photodiode with about 22 kHz as a sampling rate. Light inserted on QPD comes from a scattered light from an optically trapped particle, therefore it should change with a high frequency, but, I have just found that the measured signal changes just in some steps:
1.49414062
1.49414062
1.49414062
1.49902344
1.49902344
1.49414062
1.49902344
1.49902344
1.49902344
1.49902344
1.49902344
1.49902344
1.49902344
1.49414062
1.49902344
1.49414062
I mean some constant numbers exactly have been repeated, which is my surprise. I also have seen that resolution of my PCI should be about 12 bits so I should have 0.0024 volt resolution. Is there any idea?
Thank you,
Faha
07-23-2014 10:50 AM
Actually, your resolution should be 4.88mV based on a +-10V scale (20V peak to peak). So you are in the noise of a single bit.
07-23-2014 11:04 AM
Yes! I have just seen that by changing the minimum and maximum in data aquering, I have the real voltages. I didnt know it that I can change the resolution by Min and Max voltage and I worried I can not use this PCI. So it is solved!
Thank you,
Faha