10-04-2011 01:49 PM
I have a NI USB-6212 DAQ and am having trouble with the output signal/voltage. I have tried to do a self calibration but it always fails. I ran the DAQ Diagnostic Utility and got results that agree with what I was thinking but am not sure where to go from here. Any help would be great. I will post my Utility results below. When I* try to have it do the self-calibration it fails and won't continue so the results shown below are with that step skipped so that I could show the results.
10/4/2011 12:34:15 PM
Results saved to: no location selected (results not saved)
Selected Device: Dev1
Device Type: USB-6212 (Mass Termination)
Serial Number: 1428EF1
Device Support: (PASS)
NI-DAQmx Version: 8.9 (PASS)
Device Reset: (PASS)
The device was last self-calibrated on: 4/9/2009 7:05:09 AM
Device calibration skipped.
-------------------- Testing Counters... --------------------
Testing CTR Buffered Pulse Generation... (PASS)
Testing CTR Single Pulse Generation... (PASS)
-------------------- Testing Analog Input... --------------------
Testing AI Ground Reference... (PASS)
Testing AI Voltage Reference... (PASS)
Testing AI Finite Sample Clock... (PASS)
Testing AI Finite Sample Clock Frequency... (PASS)
-------------------- Testing Analog Output... --------------------
Testing AO Finite Sample Clock... (PASS)
Testing AO Finite Sample Clock Frequency... (PASS)
Testing AO Single Point... (FAIL)
************ Test FAILED for the following reason ************
This test failed to output 0.000000 Volts on Dev1/ao0. The actual value written was 1.706730.
Testing AO Accuracy... (FAIL)
This test failed to generate a voltage waveform with an error of less than 0.076208 Volts on Dev1/ao0. The actual average error was 4.083570 Volts.
-------------------- Testing Digital IO... --------------------
Testing DIO Lines... (PASS)
Testing DIO Ports... (PASS)
----- DIAGNOSTIC TEST COMPLETE -----
Not sure where to go from here. Any help?
10-17-2018 01:39 PM
I'm running into a similar issue. Ao0 is pinned at 1.5 V (unless the single point output is set to a larger value) and the self-calibration fails. Did you end up finding a solution?
10-18-2018 06:06 PM
Could you create a new post for this issue? Since this is an older post it is better to make newer posts even though it is a similar issue. You'll get better help this way and more visibility.