I have this same problem, and in the past have used the same solution - setting the frequency slightly higher. On a USB-6251, I used a 1 PPS coming from GPS to retrigger a 10 pulse finite train at 10.08 Hz. This was used to trigger a camera, and we just worked the 10.08 into the post processing.
Now I am revisiting the problem (now with a usb 6211), and since it has been a few years, I'm wondering if this issue has been solved in a better way. In my current application, the finite pulses are tied to analog sampling, and a fixed sampling frequency is important so that I can do some spectral analysis without interpolating to a fixed time interval in post processing.
Now I am using a 1 PPS GPS line to start off a finite pulse train of 100, ideally at 100 Hz. In this case, precise absolute time is also important (experiments over several hours will be synchronized in post-processing to a GPS-IMU for airborne sensors), which is why I am retriggering with the GPS pulse every second.
This seems like it would be an issue affecting many and would be solved by NI by now. Maybe it has and I just haven't found it. I'm sure there are multiple applications where a fixed rate counter must be resynchronized periodically to prevent long term drift. I look forward to any insight or replies by others in the same situation.
Thanks,
Ben
09-16-2009 09:39 AM - edited 09-16-2009 09:41 AM
Hi Ben,
It sounds like you guys have observed that you cannot re-trigger the pulse output while the previous generation is still taking place. The issue is that the pulse generation must complete all of the high and low ticks before it is re-armed for another trigger. With your current hardware, the workaround that you have found (generating a slightly faster pulse train so that it is guaranteed to complete between triggers) is unfortunately the only solution that I can think of as well.
Having said this, I believe we do have a solution available using our newly released X Series DAQ devices (63xx). The key difference is that we now have the option of cutting off the unnecessary low duty cycle of the last pulse so we don't have to be actively generating up to the point where the next trigger takes place. The line will stay low after the generation is complete so the result will look something like this:
Here are two ways to accomplish what you need using an X series (new features are in Bold😞
1. Finite Implicit Buffered Pulse Train Generation -- A bit of a mouthful, but you can read more about it in the X Series User Manual. Essentially the way this works is to give an array of high and low ticks which the counter will sequentially output. The last pulse in your generation should have a two sample (the minimum) low time--as soon as this is complete the trigger will be re-armed. A couple other related facts about the X Series Counters:
2. All Subsystems are Retriggerable on X Series -- So, another option would be to configure a Hardware-Timed Digital Output to re-trigger off of the reference clock. Using the same idea mentioned above, we define a digital waveform of high and low values that will correspond to a pulse train that ends on 0, but does not completely generate the last duty cycle up to the point where the trigger occurs.
There should also be a way to do something similar on an M Series device that supports correlated Digital I/O (622x, 625x, 628x)--it's a bit tricky but here's what I would try:
-Create a Retriggerable Finite Pulse Generation that's a multiple of the frequency that you want to generate. This will be the "timebase" for a digital output task. As an example, if you want a 100 Hz signal, you could generate a 1 kHz signal from the counter.
-The number of "timebase" pulses to generate should be enough such that we have entered the low period of the last pulse that we will be generating, but should be less than what would bring us up to the next trigger signal. In the above example, if pulses were coming in every 1 second you should generate 996 timebase ticks (a 100 Hz signal will be 5 high and 5 low ticks of the 1 kHz "timebase").
-Build a digital waveform that alternates between high and low ticks at the desired rate. Using the above example, we would alternate between generating 5 high and 5 low ticks. The last sample (996th) will be a low value so that the digital line will remain low until the next sample clock edge is received.
-Configure a continuous digital output task, using the counter's internal output as a sample clock. Write the waveform to the task before starting (regeneration is enabled by default, so you only have to write once).
Unfortunately, the USB 6211 does not support correlated digital I/O, so you will have to use another board anyway. If you are purchasing new hardware, I would strongly recommend the X series if PCIe or PXIe is an option (they are priced the same as the equivalent M series boards). The M series solution would require both counters and the digital output subsystems, whereas the X series can do the same task with a single counter.
I hope this is helpful, let me know if you would like any clarification.