I need to convert a LabView App written in LV7.1 under WinXP to LV2012 under Win7 (the LV Version is 32-bit, but the OS is 64-bit).
One of the devices I need to control is a 96-line DAQPad-6508, for 7 ports digital output and 5 ports digital input.
I have DAQmx 9.7.5 installed, but it doesn't recognize the DAQPad-6508:
MAX just shows 'USB\VID_3923&PID_1750\011F6868' and no driver;
the device won't show up either if I call the Property Node 'DeviceNames' of DAQmx System.
If I try to install Traditional DAQ (downloaded TDAQ750) I get
'This version of Traditional NI-DAQ not supported on 64-bit Windows'.
What do I need to do to use DAQPad-6508 with my system?
Lars Sözüer
04-29-2014 11:03 AM - last edited on 04-26-2024 05:55 PM by Content Cleaner
Lars,
Unfortunately the DAQPad-6508 is an old device that is only supported by Traditional DAQ and not DAQmx. Traditional DAQ is a legacy driver that does not support 64-bit operating systems.
Your best options are either to: