06-07-2021 10:43 AM
Hi!
I tried to acquire an analog signal in edge triggering mode using myDAQ....but was getting the error....The same thing i tried on USB-6009....using PFI0 as the trigger source and it worked well. The program is attached here with this message. Have a look and share ur views. I want to know if myDAQ can be used to acquire nalog input signal in digital triggering mode
Thanks & Regards!
Manish Vohra
Solved! Go to Solution.
06-07-2021 02:08 PM
If the DAQmx driver tells you that something is not possible then this will always be correct.
The myDAQ manual does support this as nothing is mentioned about trigger modes for AI.
Regards, Jens
06-08-2021 05:53 AM
Hi i have come across this article....which says hardware triggering is not possible in myDAQ....this is what i interpreted from it....but i would like if u an also have a view on this & let me know ur understanding of this article....
myDAQ Errors -200077 and -200452 Using Timing and Trigger VIs - National Instruments (ni.com)
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019LyJSAU&l=en-IN
Thanks & Regards!
Manish Vohra
06-08-2021 06:22 AM
I have nothing to add. As the KB states, the myDAQ simply cannot handle a digital trigger for the start of an analog task.
Regards, Jens
06-15-2021 01:13 AM
DIGITAL INPUTS (Triggering Edge) CAN NOT BE READ AS ANALOG INPUT TASK.
if you want to see in terms physical quantity like speed measurement,
For Example, Find the Number Digital pulses (either Low to High or High to Low) per second with the help of optical Encoder external circuit.
Then Convert them as an RPM. you will get Speed in terms of RPM
