Have you wired the error cluster to an error indicator ( or error display)? Or can you place a probe on the error line (which should be ) in front of sending the "Configure Digitizer Parameter" command. I suspect that an error has occurred somewhere and prevents that particular configuration from taking place.
As I understand, some m\commands do work, which meand the method used to communicate with the instrument (VISA or serial) works..
Also look at the instrument when sending the command(s) to see if an error code pops-up. There may be an out-of-bound configuration setting which is sent and the instrument ignores the command.
RayR
03-09-2007 03:20 AM
03-09-2007 08:00 AM
"Note that the front panel sample interval and buffer size settings are independent of the sample interval and buffer size that you program over the GPIB. This is because you can qualify measurement triggers over the GPIB, which makes the GPIB measurements independent of the front panel measurements. Refer to chapter 8 for more information about GPIB measurements."