Hi All,

 

I am new in LabVIEW.

Recently, I make program to read measurement result from Power Meter of VDI PM5B.

The equipment quite uniqe because the the results are in ASCII and binary.

The response from the PM5B has a six byte message format. Byte 1 is an ASCII character. Byte 2 can be an ASCII character or straight binary depending on Byte 1. Bytes 4 through 6 are straight binary. In the returned response, Bytes 2 and 3 together constitute the data in 16 bit integer format (short integer). These two bytes should be unpacked as a 16-bit integer. In addition, these are 16-bit 2's complement numbers, which will need to be converted from 16 bit 2's complement to decimal.

How to convert the 16 bit 2's complement to decimal?

 

I attach my program. it is not work well, because the result is diference with the reading of the power meter.
I also attach the explanation from the manual.

Hopefully, you can help me.

Thank you.

Manual of PM5B.jpg

 

 

0 Kudos
Message 1 of 14
(2,157 Views)

Hi kurnwin,

 

when you want to parse 6 bytes/chars then you should do that:

Maybe you need to insert a "ToI16" conversion after the JoinNumbers function, I'm not quite clear on all those "2's complement numbers"…

(There exists a lot of other ways to do that conversion from string to numeric values, like UnflattenFromString…)

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 2 of 14
(2,098 Views)

Dear GerdW,

 

Thank you for your reply.

I see the picture of your program. Would you like to give me the VI version of the program?

Or maybe you can mention the name all drivers in the program that you have made.

I try to edit my program based on your suggestion, but I do not know the name of each function/driver.

 

Thank you.

 

Best regards,

 

Windi

0 Kudos
Message 3 of 14
(2,073 Views)

Hi kurnwin,

 

I didn't save the Vi because of its simplicity.

There is a StringSubset and I already named JoinNumbers, the other functions are really very basic stuff…

 

Which items don't you know?

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 4 of 14
(2,071 Views)

Hi GerdW,

I laready found all the drivers. Howerver, the measurement result in labview is still different with power meter display (the right one).
Only byte 3 are continuously change from the measurement. Sometime byte 4 is changed. The others byte did not change.
Here, I attach the front panel result and the program.
The dislay in labview is -0.0401638, Hower the right reading id power meter' display is -0.3438. The both display should be same.

 

Thank you.

 

Best regards,

 

Windi

kurnwin9_0-1638890275319.png

 

0 Kudos
Message 5 of 14
(2,051 Views)

Hi kurnwin,

 


@kurnwin9 wrote:

The dislay in labview is -0.0401638, Hower the right reading id power meter' display is -0.3438. The both display should be same.


A value of -0.3438 is not possible with your current conversion formula!

0.3438*59576/0.2~=102411

This value would be out of range for an I16 value…

 

Please check your formula!

Which measurement range do you use? Which factor does apply for that range? Please doublecheck each setting!

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 6 of 14
(2,038 Views)

Hi GerdW,

 

Do you mean the value is out of range of I16 because the formula is wrong?

I think the formula aready right. I use the measurement range 4. So the multiplier is 200.E-3 (0.2), and I did not use the calibration factor or it is 0.

The formula in the manual  is: 

reading = countvalue*2.*rangemax(rangeval)/59576.

It is stated that Bytes 2 and 3 together constitute the data in 16 bit integer format (short integer). However Byte 2 never change.

 

 

 

Best regards,

 

Windi

 

kurnwin9_0-1638955137888.png

 

0 Kudos
Message 7 of 14
(2,027 Views)

@kurnwin9 wrote:

 

It is stated that Bytes 2 and 3 together constitute the data in 16 bit integer format (short integer). However Byte 2 never change.


That's because something is fishy. According to the document the first byte should be a 'D'. This is ASCII code 68 (or 0x44). That seems to be your byte 2. So one possible explanation:

 

Your response protocol is in fact more complex and you failed to post that section of the specification. It may contain an extra character in front of the message. 06 as ASCII code is very unusual as it corresponds to the <ACK> command, but hey your data portion seems to consist of 6 bytes. Could it be that that is not coincidental?

 

Since you fail to show the entire string that is received currently, we can not really tell you what is the proper configuration.

 

Either there is also a character 13 at the end of the received string, then you should use the VISA Configure Serial Port and change the termination character to character 0xD in hexadecimal or 13 in decimal.

 

If there is a value 10 at the end of the message (would be strange considering you send a 13 at the end of your command), you would also use the VISA Configure Serial Port and leave its termination character settings.

 

If there is no 13 or 10 at the end of the received byte stream you should also use VISA Configure Serial Port but disable the termination character handling and then first read one byte and treat it as the number of bytes that you need to read.

 

All this is highly speculative though and based on the data observed in your front panel and the lack of any further detailed documentation of the full protocol frames.

 

 

Rolf Kalbermatter
My Blog
0 Kudos
Message 8 of 14
(2,007 Views)

Hi Rolf,

 

Yes, you are right. There is a different between the explanation in manual with the fact.

The respond from the power meter: Byte 1 is always 6  <ACK>, and byte 2 is always 68 (D).

Every send a command to the power meter, it should be terminatied with a carriage return (x0d) as expalined in the manual.

There is not character 13 at the end of the received string.

I attch the reading in string.

Therefore, I confuse which are of the bytes that contains 16 bit integer format for the reading of power meter.

kurnwin9_0-1638964828938.png

 

kurnwin9_0-1638964409096.png

 

 

Best regrads,

 

Windi

0 Kudos
Message 9 of 14
(2,004 Views)
Solution
Accepted by kurnwin9

Hi kurnwin,

 


@kurnwin9 wrote:

The respond from the power meter: Byte 1 is always 6  <ACK>, and byte 2 is always 68 (D).

Therefore, I confuse which are of the bytes that contains 16 bit integer format for the reading of power meter.


Then you should try to use Byte 3&4…

 

With the values from your last image you get:

3*256+174 = 942

942 * 2.0 * 0.2 / 59576 = 0.00632469

Does that make sense to you?

Best regards,