11-03-2022 07:59 AM
It seems that all 0~30 addresses are used, but actually there is only one device.
I don't know why.
11-03-2022 09:53 AM
What GPIB card do you use?
Worth resetting the NI MAX database.
11-03-2022 11:23 AM
Looks like your device uses Secondary Addresses. While uncommon, it is a valid GPIB setup. Can you supply more information on the device you are trying to communicate with?
11-04-2022 12:23 AM - edited 11-04-2022 12:45 AM
Thank you for your reply.
The GPIB card is Keysight-82357B, and the device is Keithley 2410.
I've encountered the same situation twice in the past ten years. The last time is NI-PCI GPIB cards + Kepco power supplies.
11-04-2022 12:35 AM - edited 11-04-2022 12:44 AM
Thank you for your reply.
I never know the trigger condition of the secondary address. Actually, only GPIB0::24::10 INSTR is valid.
The equipment combination is keysight-823578 + Keithey 2410.
It seems that GPIB card is constantly looking for device address...
11-04-2022 12:06 PM
You should just send and receive commands from "GPIB0::24::Instr" in your case. Ignore the secondary addresses with all Keithley instruments.
11-04-2022 10:14 PM
Thank you.
I think you are right.
The problem is that the list of device addresses is too long to find a right one if there are more instruments, so I want to clear it.
Do you have any suggestions?