4 wire sensing erratic behavior (LabView & Keithley 2400)

I am currently attempting to take a 4 wire measurement sourcing current with a Keithley 2400. However I keep having very erratic measurements. I am currently measuring a 150.5nm Silver thin film. I am worried that the impedance of the device may not be high enough because I keep reaching compliance or the current is leaking out of the probes. When it is in the 2 wire mode the program produces expected values, but in the 4 wire it is difficult to obtain a steady reading. Please help.

 

0 Kudos
Message 1 of 3
(4,359 Views)
0 Kudos
Message 2 of 3
(4,345 Views)

Hi,

 

Are you on the 4 wire Sensing Mode? Get there by pressing " Config" >> "Ohm (symbol) >> "Sense Mode" >> 4Wire.

 

Make a hand measurement before letting LabVIEW take over. Is the arrangement of probes right? 

 

Keithley has good support so I would suggest giving them a call..

 

Regards,

Gautham

 

 