Callender-Van Dusen Equation

Hi

cn some one help me to slove this equation.
 
where R0 is the resistance which I read at runtime, 

Temperature below 0 degree Celcius.
              RBaseT = R0[1 + A X T + B X Tpow2 + C X Tpow3 X (T - 100 degree celsius)]

Temperature above 0 degree Celcius.
              RBaseT = R0[1 + A X T + B X Tpow2]

where

RBaseT = RTD resistance at temperature T
R0 = RTD nominal resistance at 0 degree celcius
A,B,C are coefficients given

A = 3.18 X 10pow-3
B = -6.02 X 10pow-7
C = -6.0 X 10pow-12
Thanks in advance.
0 Kudos
Message 1 of 6
(8,777 Views)
Hi Sujeep,
 
I know that this equation is used in our FP-RTD-122 modules, but I am not sure why you need this equation solved.  Are you trying to double check the modules output against the equation?
Regards,

Michael
Applications Engineering
National Instruments
0 Kudos
Message 2 of 6
(8,766 Views)

Hello Mr.Michael

         I am reading the resistance value from fieldpoint and converting into temperature by using the formula

          (2.5641 X Resistance value) + 16.59 = Temp in K.        i.e., y=mx+c.

        But this is not accurate value. so I want to make use of the formula provided by NI. Kindly look into the attachment.

Thanks.

 


 

0 Kudos
Message 3 of 6
(8,751 Views)
This calculation can be made quite easily with the use of a formula node.  The formula node will take in all of the inputs for this equation and will give you a result.  I have attached an example that implements your equation.  I hope that it helps.
Brian Coalson

Software Engineer
National Instruments
0 Kudos
Message 4 of 6
(8,707 Views)

Hello Sir,

Thanks for ur reply. But I could not view the attachment becz i am using LabVIEW8.0.

Thanks.

 

0 Kudos
Message 5 of 6
(8,679 Views)

Hi Sujeep,

Here is the VI that Brian wrote but for LabVIEW 8.0.  I hope this helps.

Regards,

Michael
Applications Engineering
National Instruments
0 Kudos
Message 6 of 6