02-22-2010 07:30 PM
I am currently using a Keithley 2410 Sourcemeter for my lab and I use it in my experiments as a voltage source for an RPA probe. I have written a VI for it that increases the voltage from say 0-500V at 5V increment. The speed at which it is increasing the voltage is set at its fastest at 0.01*NPLC, where NPLC = number of power line cycle, and 1*NPLC = 60Hz. The problems come when there are other "disturbances" in the circuit that causes the sourcemeter to slows down, such as from the plasma. This happens more often as I increase the voltage from another power supply that controls the plasma and every time it does, there seems to be a pause/hiccup from the sourcemeter. Are there any other power supplies or voltage sources that can provide a faster increase than 0.01*NPLC?
02-23-2010 04:00 PM
Hi MaxXx-
I have a couple of questions to clarify what you need.
1. Do you absolutely need to go to 500V? The highest we can source is 80 V by cascading our PXI-4110s.
2. By 0.01*NPLC, do you mean 0.6 Hz (60 Hz*0.01)? If you do mean 0.6 Hz, we should be able to easily beat that for speed of a voltage sweep.
Let me know if 80V max is sufficient. I am happy to help with any further questions you might have.
02-24-2010 09:12 AM
MaxXx,
There is an application note on the Keithley web site on how to optimise the throughput of the sourcemeter:
http://www.keithley.com/data?asset=11641
The nplc has to do with the reading measurment time - do you need to measure or just source?
02-24-2010 03:25 PM
National Instruments actually has an SMU that can source/sink 100V on a single channel.
http://www.ni.com/powersupplies/