11-08-2019 05:29 AM
Hi guys,
Just asking if there is a way to export the external calibration of PXIe-4081 DMM? I tried NI-MAX but it doesn't show on the device list. Thanks in advance.
11-08-2019 10:57 AM
There isn't such functionality.
Can you explain more about your use case so we can get some context of why you want such a thing?
11-11-2019 12:54 AM
Hi Marcos,
I planned to reformat one of our machine and if possible i want to backup the calibration file/data of PXIe-4081 DMM so we don't need re-calibrate it again after reformat. Is there's a way to do this? Thanks.
11-11-2019 10:53 PM
The calibration constants are stored on a flash memory on the instrument. You can wipe your hard drive clean and/or move the instrument to another system, calibration constants are preserved.
Do note that in order to meet the documented specifications for the device you need to run self-calibration in the last 24 hours after the device has been powered up for 60 minutes on the system it will be used in.
See http://www.ni.com/pdf/manuals/375254c.pdf