07-09-2012 03:37 PM
07-09-2012 03:51 PM
Does it need to be automated? A cheapo SPST across the input will do the job...
My bet is the input impedance of those terminals goes high so you don't accidentally short something out if you were to reconnect it to something else.
07-09-2012 05:00 PM
I think SnowMule hit the nail on the head.
Another option would be to set the 9219 to voltage mode and put a resistor across the input terminals. Then scale the voltage reading to the associated current using Ohm's Law.
-AK2DM
07-09-2012 06:19 PM
Then the meter would be off since the additional resistor would be in the current loop. You'd need an extremely low value resistor, and read extremely small voltages across it.
07-10-2012 07:42 AM
Where is it write that the output is high impedence when the power is off? I did'nt find the information in the user manual...
07-10-2012 07:46 AM
http://www.ni.com/pdf/manuals/374473e.pdf
Input impedance
Voltage and Digital In modes
(±60 V, ±15 V, ±4 V) .................. 1 MΩ
Current mode .............................. < 40 Ω
All other modes .......................... >1 GΩ
Is it possible to set a hardware configuration to stay in 'current mode' when the card power is off???
07-10-2012 10:24 AM
SnowMule:
Current loops can have multiple readout devices in series as long as the total voltage drop is below the compliance voltage of the source power supply.
-AK2DM
01-26-2018 03:11 PM
Use a power diode across the two terminals of you cDAQ device. take special care to keep the orientation of the diode in the direction of the loop. when you device is on current will flow through your device and when it is off it will flow through the diode keeping the loop intact.
01-29-2018 02:16 AM
It depends on the specification of the 4-20mA source. Most devices I know can drive a 500-750 Ohm resistance. The resistance of the 9219 input is < 40 Ohm. So it is low but not a fixed value.
With a normal industrial 4-20 mA current source you can add an additional resistor in the loop of 250 Ohm and measure the voltage across this resistor (0-5V).
For the current loops I know this is common practice. e.g. to read communication data (HART) from the device.
Kees