10-30-2008 01:19 AM
When I use the example with NI-SYNC called Generate DDS Clock and Route.vi and select DDS Clock Source Terminal: ClkIn, and Destination Terminal: ClkOut, the card does not generate an output clock. If I then click stop, I'm given the error message:
Error -1074118606 occured at niSync Connect Clock Terminals.vi
Driver Status: (Hex 0xBFFA4032) The specified source terminal is invalid for this operation.
If I change the DDS Clock Source Terminal to one of the other sources, I get a signal on ClkOut, but it is not phase locked to my external 10MHz reference oscillator.
How do I generate a DDS clock signal which is phase locked to an external 10MHz reference oscillator connected to ClkIn?
John Bosshard
Solved! Go to Solution.
10-30-2008 06:17 PM
Hi John,
You need to use two LabVIEW examples for this task. Run the Check Clk10 & Route Clock.vi first to route ClkIn to the backplane PXI_Clk10. Then run the Generate DDS Clock and Route.vi to route the PXI_Clk10 to ClkOut. This will work as long as your signal meets the backplane 10MHz clock specifications.
Thank you,
10-30-2008 06:25 PM
Even when run by itself, "Check Clk10 & Route Clock.vi" fails. Upon running it, I get "Error -1074118606 occured at niSync Connect Clock Terminals.vi"
Possible reason(s):
"Driver Status: (Hex 0xBFFA4032) The specified source terminal is invalid for this operation."
So--essentially the same error as above.
10-30-2008 11:38 PM
Solved--PXI-1033 S1 backplane jumper needed to be moved.
John Bosshard
10-31-2008 10:03 AM
Hi John,
Thank you for letting me know and I am glad everything is working now.