Hi rhenson,
It seems unlikely that this problem is related to the operating system. I suspect that there is either a setting in the software or there is a hardware issue.
The first thing I would check is the Resource Manager Delay. This delay specifies how long Resman, the resource manager, will wait before it configures all of your instruments. This setting can be set in Measurement and Automation Explorer (MAX). Right click on your VXI System, select Properties, and select the Settings tab. By default, the Resource Manger Delay should be set to 5 seconds. If the delay is set to a higher number, I suggest you use a lower Resource Manager Delay.
It is also possible that this is a hardware issue. To determine if your VXI instrument or the MXI-3 system is
causing the problem, I suggest you test your VXI system without this VXI instrument. Swapping in another instrument or another MXI-3 kit would also be a good idea.
If none of this works, we'll need more information about this system before we can continue. Where exactly does the system hang? Watch Resman to find out what step is taking a long time.
Regards,
David Shatila
Applications Engineer
National Instruments