05-01-2018 03:53 PM
Hi,
I am running the NI USRP Simple Streaming project (without any modifications) on a 2944R and noticing a large DC offset and I/Q imbalance. The screen captures below show
outputs without anything connected to the RF 0 receive channel.
These vary over time. For example, at the start of one run the DC offset was at -60 dBm. After about ten minutes it sank down to -80 dBm and after another ten minutes it crept back to -70 dBm and higher.
How do I turn on impairments correction in the Simple Streaming project? There is no
control on the front panel. If it is already turned on by default, it isn't producing great results.
Regards,
Ivan
05-03-2018 09:47 AM
The corrections are enabled by default in the sample project, and I believe you are seeing good results. I would say LO Feedthrough at -70 dBm is good for this device. However, you have some additional options.
There is a self-cal utility for Rx DC Offset (LO Feedthrough). I recommend the following steps
If this is still unacceptable, you can
05-03-2018 12:30 PM - edited 05-03-2018 12:32 PM
Hi Ivan,
Thanks for reaching out! Unfortunately, the USRP is an uncalibrated/unwarranted device, so this behavior is expected and is somewhat common. The I and Q signals may go through separate signal paths that could cause different amounts of imbalance in the final reading. It also may be due to temperature fluctuations when starting acquisition/the device.
There are a couple steps we can take to potentially compensate for this, though:
Also, it would be good to know how you are planning on using this device/what measurements you're trying to take. The IQ imbalance may not be ultimately critical in terms of some tests.
Thanks,
Sunayna
07-25-2018 01:32 AM
hi, did you solve this problem? I am also have this problem,and I have no idea how can I solve it
07-25-2018 03:41 PM
As psisterhen suggested using the LO offset to shift the DC bias out of the filter passband can cut the DC bias quite substantially. The I/Q imbalance is also diminished. However this only works if you reduce the bandwidth. It doesn't work if you need to use the full sample rate of 200 MS/s.
Calibration would probably help but I haven't had time to explore this option yet.
07-26-2018 01:50 AM
Thank you !