03-25-2008 12:19 AM
We have a test system which uses NI PXI 2570 (40 Channel SPDT). We control some data and lines including the DUT ON/OFF with these relays. Before running each test case, I would like to have a resetting of relays (using Disconnect all) except the relay where I connected the DUT ON/OFF. Could anyone suggest a best method for this?
Thanks in advance,
Sanad MM
03-25-2008 08:15 AM
03-26-2008 11:38 AM
Hi
We are using NI Switch Executive Version 2.1.1. I have tried with “connect and disconnect” option. Now I have one question. We are using PXI 2570 (40 Channel SPDT). I have attached the code which I am using. When I disconnect NO-COM connection, it is not physically disconnecting the connections. If we want to physically disconnect we have to switch to NC-COM (as in picture). Is this one the expected behavior?
Sanad MM
03-26-2008 11:56 AM
Hi Sanad MM,
As you note, the command to disconnect NO-COM does not physically connect NC-COM. This is expected behavior.
Disconnect does not activate the relay until the ensuing Connect is executed. Disconnecting NO-COM informs the software so that connecting to NC-COM can occur without error.
Hope this helps!
Chad Erickson
Switch Product Support Engineer
NI - USA
03-26-2008 12:01 PM
Addendum: This is the correct behavior for SPDT switch modules. These switch modules include the PXI-2548, PXI-2549, PXI-2558, PXI-2559, PXI-2566, PXI-2570, PXI-2598, and PXI-2599, as well as the SCXI-1160, SCXI-1161, and SCXI-1166.
Chad Erickson
NI - USA