10-20-2015 02:41 PM
When I look at the PXI 2521 specification, it says the PXI input trigger lines 0-7 are supported. I am trying to route a Data Marker Event from my function generator (PXI 5402) to these switches. But when I view the Device Routes for the PXI 2521 in NI-MAX, it doesn't show these lines as available as a source; it only shows TrigIn and AdvComplete.
What am I doing wrong here?
10-21-2015 02:48 PM
Hello,
My name is Hyrum, from Applications Engineering, the reason why you only see two triggers on NI MAX, is because the switch itself only has a Trigger in and a Trigger out. Those then have sources that you configure in LabVIEW or other programming language. I hope you find this useful.
Regards,
10-21-2015 04:12 PM
jvavra,
Don't think of sources as places where we can output a signal. Think of sources as terminals on the FPGA of the switch. In that behalf, TrigIn and AdvComplete are the only sources on these devices.
That doesn't mean that we can't output a signal from the device to the PXI_TrigX lines. It just means that we output a signal from the AdvComplete terminal of the FPGA and route it to the PXI_TrigX Line.
Examples
Output trigger onto PXI_Trig0:
You can route AdvComplete to PXI_Trig0. This allows a signal to be sent from the FPGA to the AdvComplete terminal, which is then sent to PXI_Trig0 on the PXI Backplane.
Read trigger from PXI_Trig1:
You can route TrigIn to PXI_Trig1. This allows a signal to be sent from the PXI Backplane to PXI_Trig1 on the PXI Backplane, which is then sent to TrigIn on the FPGA.
10-21-2015 04:23 PM
jvavra,
Also, the software doesn't really look like that table anyways and is in fact much simpler. You simply set up a scan-list and configure trigger inputs.
For example:
If you wanted your PXI 5402 to output a trigger and have it reach your switch, you would: