You should be able to obtain much better results by increasing the number of samples to read. A buffer size of 1000 with a sample rate of 100kHz means that SignalExpress must sustain a loop rate of 10ms, which may be more than the application can handle if you choose to view the data while logging is in progress. If you change the number of samples to 20,000 for example, the loop rate will increase to 200 ms, which should be sufficient.
I hope it helps!
Regards
05-23-2007 08:05 PM
09-03-2007 07:36 AM
Dear Phil,
The error it gives is that when exporting to Excel is that the rows are too much, the error says that we must go in the options and reduce the number of samples, I went there and reduced the sample number but still the same error. I do not have a latest excel, only 2003. Any ideas?
Munir.
09-04-2007 09:43 AM
Hi Munir, are you using SignalExpress 2.5? I was able to get this to work correctly using that version. I'll check using SE 2.0 in case that is what you are using...
Phil
09-04-2007 10:52 AM
Hi Phil J,
I am using LabVIEW SE ver.2.5.
Munir