The NI 5411 does not natively support finite generation mode. The best workaround is to configure the device in Arbitrary Sequence mode and download two waveforms. If no trigger is used (trigger source = immediate)the first waveform would represent the pulse you want to generate and the second would be a dc waveform with the minimum number of samples required by the device (256 for the NI 5411). Configure the device in stepped trigger mode.
With this setup, the NI 5411 will now generate your pulse ones and than generate the dc signal until the generation is stopped.