10-23-2007 01:51 PM
10-23-2007 04:42 PM
10-24-2007 02:19 PM
Hello Rowlini,
I am assuming that you are Right-Clicking on Traditional NI-DAQ under Devices and Interfaces in Measurement & Automation Explorer (MAX) and selecting “Reset Driver for Traditional NI-DAQ” to reset the driver. What do you mean that all 5 channels are not valid? Are they not valid because you are getting error -10007? Are you getting this error through a test panel?
What test panel are using to test your DAQ board? Are you using the test panel for the device or the test panel for the virtual channel? I would use the test panel for the device first and make sure that you are acquiring the raw data correctly. Then open the test panel under virtual channel to verify that all the settings are correct.
This error might be created because the settings are not set correctly in the virtual channel. I would suggest double checking those values. Make sure that your virtual channel is using the correct analog input mode. (Differential or Referenced Single Ended) Do you have the correct channels selected if you are using differential mode?
10-24-2007 04:08 PM
10-25-2007 10:30 AM
Hello Rowlini,
Normally, you would prefer to measure your sensors using differential mode. This mode will try to eliminate as much noise as possible. The problem with this method is you have effectively reduced the amount of channels from 16 to 8. The TCOX modules will actually measure the thermocouple using a differential amplifier circuit and measure the CJC. The differential measurement is then passed to DAQ board on AI (X) and the CJC measurement is passed to the DAQ on AI (X+8). There is CJC data is always passed to the DAQ board on AI (X+8) with these modules. There is a great circuit diagram explaining how the modules operate on page 5 of the SCC-TC Series Thermocouple Input Modules User Guide.
You would then use Reference Single Ended mode to measure
the Thermocouple measurements on the DAQ board.
(They were differentially measured on the module) There is a great explanation of the different
measurement modes in chapter 4 of the AT E Series User Manual. The biases resistors are not necessary;
however it might become useful if there is a lot of noise on your signal.