Signal Conditioning

cancel
Showing results for 
Search instead for 
Did you mean: 

SCXI-1122 w/SCXI-1322 signal noise

The SCXI-1122 with W1 at (factory setting) and W2 (pullup). 

In using a combination of a thermocouple and 0-10 / 0-5 vdc at the SCXI-1322 wired inputs on CH0-CH7. 

All are set to differential, 4khz and looking at the physical input it seems stable, at least not to match the data seen in attachment.

It appears as a sensitivity issue in this case, rather than noise.  I'm unfamiliar with the SCXI and using multiple type inputs on the same module. 

 

0 Kudos
Message 1 of 3
(2,399 Views)

the attached signal display (noise) is present on all channels when no change to the voltage input signals are present present.

I mis-spoke about the 4-20ma input.  There is a 250 ohm  at the device and NIMax is set to read 1-1-5vdc.

when the level of the signal changes, the input appears to change appropriately without the noise amplitude, as if common is missing.

Thank you all for any suggestions!

0 Kudos
Message 2 of 3
(2,295 Views)

How do i adjust signal sensitivity, in NiMAX?

Verified that noise is not sourced from field wiring by connecting a battery box w/short leads to the input channel.

 

thanks 

0 Kudos
Message 3 of 3
(2,239 Views)