AISENSE is ground referenced on the SC-2043-SG. In order to avoid a ground loop, which may affect your measurements, all analog input signals MUST be floating. This pin is the reference node for the conditioned strain gauge bridge signals. AISENSE is directly connected to the excitation supply return and is earth-grounded. This makes AISENSE a low impedance reference. Therefore, all signals referenced to it MUST be floating.
Configure any unused channels for full-bridge connections and short their input screw terminals (CHn±) to their excitation return screw terminals (-EXn). If you are using an external excitation voltage, the -EX input signal is connected to DGND on the SC-2043-SG. Because this lowimpedance connection can cause a ground loop, which may affect your measurements, your excitation voltage source MUST be floating. If you are using an external + 5V power supply, the ground input signal is directly connected to DGND on the SC-2043-SG. In order to avoid a ground loop, which may affect your measurements, your +5 V power supply MUST be floating. Remember that your DAQ board must be configured for NRSE operation and all analog inputs are referenced to AISENSE.
Hope this helps.
Matthieu G.
National Instruments