The SCXI-1520 uses the SCXI-1314 Module which does not have cold-junction(CJ)compensation capability - meaning it can not compensate for either the temperature of the connection to the DAQ system, nor, the additional voltage effect due to the dis-similar metal contacts of the DAQ system. This correction is absolutly necessary for accurate temperature measurement. I suggest you use a "third-party" cold junction module from a vendor such as OMEGA Engineering (www.omega.com). The CJ sensors are battery powered and reference the signal to 32 Deg F while at the same time providing standard copper leads for the DAQ system.