Online shopping is currently unavailable.
Please contact us for help or try again later.
02-16-2026 07:12 AM
The NI-9250 specification states a 50 ohm impedance AI- to chassis.
I use the device with BNC-connector and it is clear, from documentation, that AI- is the shield of the BNC-connector. But where can I find a defined connection/point/terminal of the chassis?
I have not been able to measure the claimed 50ohm anywhere.
Between where can I find the stated 50 ohm?
Solved! Go to Solution.
02-16-2026 10:01 AM
Chassis connection is indirectly referring to the GND plane of the card and electrically when inserted into the cRIO/cDAQ chassis, you will be able to measure 50ohm between the AI- and any exposed metal portion of the chassis or chassis screw.
02-26-2026 01:15 AM
No, 50 ohms can not be measured from AI- to ANY point of the NI9250 (chassie, DSUB-sockets, Earth screw etc).
It is more like 2.2 Mohms
Can anyone please explain the by NI stated 50 ohms between AI- and Chassie. Or is the NI-9250 specification wrong?
02-26-2026 03:10 AM
What do you measure (using a DMM) between both AI- after configuration while the task is running?
02-26-2026 03:16 AM
DC resistance using a DMM Fluke 789. NI-9250 switched off, nothing else than DMM is connected.
Both AI- and chassie are defined by marking on the product. Definitely not 50 ohms between them.
02-26-2026 07:01 AM
That's why I wrote, measure while configured and task is running 😉
02-26-2026 07:08 AM
Thanks. 50 ohms confirmed when the device is active/ready (power on).