A dumb question from first time user:
From the specification of SCC-SG01, the voltage input range is -100mV to 100mV. But if no signal/sensor is connected to SCC-SG01, is that equivelent to RG = infinite so that voltage input becomes 2.5V/2 = 1.25V which exceeds the limit?



0 Kudos
Message 1 of 3
(2,837 Views)

Hello,

You are correct.  The signal would be floating at 1.25 Volts, which would cause your measurement to rail at 100 mV.  The SCC-SG01 has over voltage protection up to 28 V, so 1.25 V will not damage the module.

Please see the SCC-SG Series Strain-Gauge Modules User Guide for more details at:  http://digital.ni.com/manuals.nsf/websearch/876393DF257DB5B086256E55005A51CE
 
I hope this helps.
 
Regards,
Sean C.