Hi NSWC Ben,
Welcome to the NI Discussion Forums! You pose a great question and you are right in saying that the operating instructions for the NI-9237 do not explain how to enable the shunt calibration in software.
There are two ways to enable/disable the shunt calibration:
1. In Measurement & Automation Explorer (MAX) using a strain task:
If you already have a strain task configured, you can skip
to step 6.
2. In LabVIEW:
DAQmx 8.7 has added the DAQmx Perform Shunt Calibration VI to enable/disabled shunt calibration. If you have DAQmx 8.7 installed this VI can be found on the functions palette under Measurement I/O >> NI-DAQmx >> Advanced >> Calibration >> DAQmx Perform Shunt Calibration VI.
There is an example of this in the example finder as well. Help >> Find Examples. Navigate to Hardware Input and Output >> DAQmx >> Analog Measurements >> Strain >> Cont Acq Strain Samples (with Calibration) – NI 9237 VI.
In earlier version of NI-DAQmx, the shunt calibration is enabled with DAQmx Property Node. I have attached a screenshot of how to get to the property you need in case you have an earlier version of NI-DAQmx.
04-15-2008 01:36 PM
04-16-2008 01:57 PM - edited 04-16-2008 01:59 PM
The best method is to use the DAQmx Perform Bridge Offset Nulling Calibration VI in DAQmx to remove offset errors. There is also a shipping example in LabVIEW that demonstrates how to programmatically perform a offset null offset calibration. It can be found from the NI Example finder: Browse According to Task » Hardware Input and Output » NI-DAQmx » Analog Measurements » Strain » Acq Strain Samples (with Calibration).vi. The example I mentioned in my previous post also has this functionality. When you create a Strain Task or Global Channel in MAX, you can also perform an Offset Null Calibration by clicking on the "Device" tab, and then clicking on "Calibrate".
02-17-2009 11:09 PM
Jared,
Are you aware of any problems with the DAQmx Perform Bridge Offset Nulling Calibration VI and Custom Scaling using an NI-9237 ?
Thanks
Dan
02-19-2009 08:50 AM
Hello Dan,
I have looked through our database for any reported problems and have not come accross any. Is there a particular problem that you are having?
02-19-2009 11:48 AM
Margaret,
I found a workaround to my problem through a different post:
"Why do I receive Error 200378 "Custom scale specified does not exist"?
Thanks
Dan