Error - 89133 (System: PCI-6024E, SCXI-1000, SCXI-1520, SCXI-1314)

I am trying to conduct some strain gauge measurements, but am having difficulties with my SCXI-1520. The error that I receive when I run the 'test panel' in 'Measurement & Automation Explorer' is:
Error -89133 occurred at daqmxcfg_DAQmx Start Task.vi
Possible reason(s):
Specified inversion cannot be satisfied, because the hardware does not support it.
Source Device: SC1Mod1
Source Terminal: SYNC
Destination Device: SC1Mod1
Destination Terminal: SCXI_Trig1

Any advice would be appreciated.
0 Kudos
Message 1 of 2
(2,989 Views)
Hello,

Thank you for contacting National Instruments.

This error is usually associated with a configuration problem. You will need to ensure that you have entered the correct parameters for your strain measurement. You will need to check that you have entered valid parameters for your excitation and bridge configuration. The default bridge mode is FULL so you will need to change that setting if you are using a quarter of half bridge configuration.

Regards,
Bill B
Applications Engineer
National Instruments