Signal Conditioning
Buffered DI from SCXI-1162HV using DAQ PXIe-6361
Hello, good morning
Is it possible to make a Hardware-Timed acquisition from a SCXI-1162HV in serial mode with a DAQ MIO PXIe-6361? I tried all C examples but none of them worked. It seemed to me that in serial mode the only way for DI acquisition from a SCXI-1162HV, using a DAQ PXIe-6361 is using the On Demand (timing) mode. Is it correct?
Thank you in advance for any help.
(4,043 Views)
Hey Cmnap002,
Unfortunately it is not possible to do Hardware Timed DIO with your PXIe-6361 and your SCXI-1162HV.
According to the specifications for the PXIe-6361, it has 8 Hardware Timed Inputs (1 Port) up to 10MHz. But a combination of PXIe-6361 and SCXI-1162HV will not allow you to do what you want.
I found a tutorial showing how to do buffered digital input for isolated or high-voltage applications . NI 6533 or NI 6534 device (Digital IO Boards) can be used with 1162HV to do HW Timed DI.
I hope that helps.
Felipe Flores
Applications Engineer II
National Instruments Brazil
(4,022 Views)
Accepted by topic author Cmnap002
Obrigado, Felipe.
Abraço,
(4,007 Views)
