Hi,
The bandwidth of SCXI-1125 with SCXI-1313A high-voltage terminal block for 100:1 attenuator setting and with 10 kHz filter is 2.6 kHz.
Is the bandwidth of SCXI-1125 with SCXI-1313A for 1:1 attenuator setting 2.6 kHz or 7 kHz?
Thank you.
03-08-2011 01:40 AM - edited 03-08-2011 01:41 AM
Dear Bojan,
Reading the specifications I came to the conlcusion that SCXI-1313A limits the bandwidth on SCXI-1125 to 2.6kHz with all attenuation settings.
You need to calculate with 2.6kHz anyway.
http://www.ni.com/pdf/manuals/372425c.pdf
(Page A-5)
Best regards,