From Friday, April 19th (11:00 PM CDT) through Saturday, April 20th (2:00 PM CDT), 2024, ni.com will undergo system upgrades that may result in temporary service interruption.

We appreciate your patience as we improve our online experience.

Real-Time Measurement and Control

cancel
Showing results for 
Search instead for 
Did you mean: 

cRIO 9074 with ni 9213 Thermocouple Type K Setting up

Solved!
Go to solution

Hi guys,

I’m neu with Compact rio and working with CompactRio 9074 with Ni 9213 for my Masterthesis. Ich use Thermocouple Type K to measure the temperature of my actuator. My question ist how can i set the TC type to type K. My cRIO don’t habe the DAQmx Funktion. Ist it possible to set it up in NI MAX oder have to do it in LabVIEW Code? I have already insert TC type K into the 9213 and can read the room temperature. The reading temperature is wrong, when i heat the room up. It goes not like a type K.

 

I hope someone can help me. Thank you.

0 Kudos
Message 1 of 3
(730 Views)
Solution
Accepted by topic author Ninewguytu

If you are using Scan Engine mode, right-click and select Properties to configure the thermocouple type.

ZYOng_0-1676991761302.png

 

If you are using FPGA mode, the conversion is done on the RT side. Please refer to the shipping example <LabVIEW>\examples\CompactRIO\Module Specific\NI 9213\NI 9213 Getting Started

0 Kudos
Message 2 of 3
(700 Views)

Thank you so much. control the prototype in RT mode.

0 Kudos
Message 3 of 3