05-01-2019 07:05 PM
I am trying to measure temperature reading from K type thermocouple using NI-9220 with the NI-MAX. Is that right to use NI-9220 for reading a thermocouple? I am new to this. Please advise
Thanks in advance.
05-02-2019 06:13 PM - last edited on 01-11-2024 11:17 PM by migration-bot
Hello Vasu11,
It is possible to take temperature readings from the NI-9220 device, however NI does have specific temperature input modules that would give you a much more accurate reading. The 9220 has an input range of +/-10 V, whereas a temperature input module, such as the NI-9210, has a much smaller input range (+/- 80 mV in this case). A typical K-type thermocouple's output generally ranges from -6.5 mV to 55 mV, so the smaller input range of the 9210, along with a high resolution, will give you a much more accurate reading. There are also other benefits to having a temperature specific input module such as open-thermocouple detection, cold-junction compensation, and anti-aliasing filters.
C-Series Temperature Input Module
https://www.ni.com/en-us/shop/category/temperature.html?productId=118148
Garrett W.
05-02-2019 06:34 PM
Hi Garrett,
Thanks for your reply.
I am wondering does NI-9220 has an in-built CJC?
Please advise.
Vasu
05-03-2019 09:55 AM - last edited on 01-11-2024 11:17 PM by migration-bot
Vasu,
The 9220 will not have a CJC, it is a temperature-input module specific attribute. I linked the datasheet for the 9220 below for you to review at your leisure. There are work-arounds to temperature compensation that you can look into as well, a good example is using a AD8495 thermocouple amplifier. I linked a website below that you can look into for examples.
NI-9220 Datasheet
https://www.ni.com/docs/en-US/bundle/ni-9220-specs/page/specs.html
Measuring Temperature Using Thermocouples
https://www.analog.com/en/analog-dialogue/articles/measuring-temp-using-thermocouples.html
Hopefully this helps,
Garrett W.
05-03-2019 12:47 PM
Hi Garrett,
Thank you so much for your help!!!
Best Regards
Vasu