How can I configurate the pxi4132 to the best sensitivity - 10pA to current measures in the 10uA of range?

0 Kudos
Message 1 of 4
(5,687 Views)
Solution
Accepted by topic author Seixas

According specifications manual when you are using the 10 uA range you are using the 10pA sensitivity, but the accurracy is (0,028% + 1 nA).

 

regards,

Abel Souza
Engenheiro Eletrônico
LabVIEW User since 8.5
0 Kudos
Message 2 of 4
(5,674 Views)

In your manual is wrote: "For instance, on the 10 μA current measurement range, you can typically achieve 100 pA sensitivity at up to 1000 readings per second."

 

I understand this is not true. Is true for 10uA currente measurement range, we have 900pA sensitivity...

0 Kudos
Message 3 of 4
(5,671 Views)

Dear Abel, 

How do you configure the smu PXI-4130 to work by supplying 5V with 1.5A through the CH0 or CH1 output channel using the auxiliary APS-4100 (12V output)?

 

Best regards,

 

Seixas

0 Kudos
Message 4 of 4
(2,400 Views)