02-28-2013 04:30 AM
Hi everyone!
I am using NI-4300 and i was wondering what happen if I acquire signal greater than max input range.
I have set the range +/-5 V, and +/-10 V and I dont see any differences when I acquire a 6V signal.
Thanks in advance.
02-28-2013 05:42 AM
I know many measurement devices will go slightly above the rated range. But I wouldn't expect a full volt above the 5V range (that's 20%). Are you sure there isn't any clipping? Which terminal block are you using?
02-28-2013 08:29 AM
03-07-2013 09:16 AM
Hi gomarq,
as far as the NI-4300 specifications manual is concerned:
http://digital.ni.com/manuals.nsf/websearch/173D788886E46B37862576D3005222D9
this board has the following fault protection:
Between AI+ and AI–.................±60 V
Between any AI and COM..........±60 V
Between PFI or RSVD lines
and COM.....................................±24 V
Input current during
fault conditions ................................±5 mA max/AI pin
It CANNOT support overvoltage during long periods. You have to set the board range according to the level of the signal you expect to acquire. If you set a range of 0 to 10V and you acquire 15 volts during several seconds, your board will be damaged and you may be hurt too. So, I do NOT recommend to do such thing. If your signals are over range, you will have to attenuate them before the 4300 acquisition.
I hope this helps!
Jesús
03-08-2013 05:23 AM
@gomarq wrote:
I am using TB-4300.
If i set the range -10/0 it also measure positive values.
Have you checked the ACTUAL range used (selected by the driver)?
03-08-2013 06:35 AM