Why does PXI 6224 give a channel voltage and all channels have the same value?

Solved!
Go to solution

The voltage of PXI 6224 in the real-time system is collected through veristad. I give a channel 3V voltage, and the 32 channels collected are 3V. Measured with a multimeter, only the first channel has voltage。Such as the attached picture。I don't know how to deal with him, ask for help。。。。。。。

 

0 Kudos
Message 1 of 2
(814 Views)
Solution
Accepted by topic author OTGZzz

A most common reason for such thing happening would the because the additional channels are not connected to any signal and the signal on one channel is inducing a charge on the rest of the channels. Instrument with high input impedance is not recommended to be left floating.

Please connect any unused Channel to GND to avoid this behaviour. This is similar to OPAMP where the inputs should never be left floating and inputs of any unused OPAMP in a package should be grounded.

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.