09-24-2021 11:31 AM
Hardware
Chassis PXIe-1082
Modules PXIe-4464 (BNC-input)
Question
I have a shielded twisted pair terminating in a BNC fitting and a shield wire at the data acquisition system. The senor (geophone) is not grounded and thus the shield wire must terminate at the measurement end. Where should I connect the shield wire? My thought is the shield wire should connect to the chassis (but where?).
Solved! Go to Solution.
09-25-2021 09:45 AM
There is a chassis ground screw on the back of the chassis you can use.
09-25-2021 01:01 PM
Found it - Thanks!